FLIR Duo Pro R

Hi, has anyone had any luck with this? 

I am using the Flir Duo Pro R 13mm 640x512. I can’t find the sensor settings anywhere but I believe it is the same sensor as that tau2. 

 

Typically I use Agisoft to process data but this camera is so new and as I see now even Pix4d doesn’t have a workflow for processing.

Currently the dataset I have is 500 or so images, 60m AGL, 2secs per image, moving at 4m per second. Flight lines distanced approximately 15m apart. North facing orientation, flying in a grid. Overlap is very high. Can visibly see the overlap between images.

The data set aligns and calibrates 100% in photoscan, but in Pix4d with the recommended settings I see for the Tau2 camera, it does not align thermal, or RGB or a combination of them both. I have tried on Alternative and on standard. Any ideas?

 

Hi Ina, Mark

Could this (Mark’s) data set be a basis for creating a proper camera rig,  for this camera and Pix4Dmapper?
My FLIR Duo Pro R / 13mm has still not been delivered, unfortunately.

Best
Reto

I still have a very fundamental question, regarding FLIR Duo Pro R images or other multi-sensor imagery:

Since the two sensors are combined in one camera, is it (one day) possible for Pix4Dmapper to calibrate the camera (rig) using the higher resolution RGB images instead of processing both image sets sequentially? If I understand correctly, that’s where the "camera rig concept would come into place. Is that correct?

For such cameras / camera rigs, which Pix4Dmapper process steps are done once only (for both sensors) and which steps are processed twice, once for RGB and once for IR?

Reto

 

Hi Reto, 

Yes, I have created a support ticket with Pix4D support, to query why it’s not currently aligning. I have unlimited access to this Camera and am out in the field approximately 3 times per week. We conduct surveys across the UK and are testing this camera out, so when I am on site conducting a survey, I will finish off and then put this camera up to do some test flights. 

The only issue we may see is I do not use ground control points. We survey solar farms and creating an orthomosaic and DEM is not our primary concern, this is a trial to see if it’s a viable delivarable to add to our current offering. If we can accurately and efficiently create a thermal orthomosaic of the site then we can begin adding this to our offering as a bonus feature. 

I currently have two complete data sets for two different sites, taken at 60m and 65m above ground level. They definitely have sufficient overlap because they calibrate and align 100% in Agisoft. The one site, I flew at 65m Above take off, taking an image every 2seconds, 4m/s flying speed, double grid pattern flown with drone deploy as the flight planning app - the overlap was set to 80% however due to the size of the sensor I’m not sure this actually changes anything - typically the flight plans are based off an RGB sensor for example the phantom 4 pro which is a lot bigger I believe. Regardless, I flew a double grid/cross hatch pattern. There is definitely sufficient overlap. I have stitched this set all the way to the orhtomosaic stage in AGISOFT, but here is where I am struggling to export a radiometric orthomosaic. Haven’t done one before. 

Hence I looked now at Pix4D as I know some people have had success with it previously, I am currently using a Trial version but it may expire in the next few days so I hope to be successful before then. 

 

The second data set was flown at 60m AGL, flying at 4m/s, flights were planned manually with Litchi flight planning. Images captured as Radiometric JPG every 2 seconds. Parallel flight lines, approximately 15m between each line so the overlap is definitely sufficient. The drone was oriented North (solar panels are south facing) and I was flying waypoints along the rows of panels (east to west). Again this imagery aligns perfectly in Agisoft but does not seem to align in Pix4D (this is the set I refer to in my first post above). 

I have contacted Flir Directly regarding the camera specifications and they say it is the same sensor as the Tau2 but they then sent me incorrect data - so I have escalated the enquiry further and am waiting to hear back. 

Hi Mark

Thanks a lot for the in-depth description of your tests, in a few days I will be in exactly the same situation, I guess. I’m using Pix4Dmapper for TGB mapping for a very long time and I’m confident that the 1st grade support will solve this issue too. If you want to drop me your e-mail address, we can exchange test results later as well. Mine is reto.weiss@pixmap.ch.

Regarding communication with FLIR, I can confirm that this is a rather sad story…

 

Hi Mark,

 

I have just sent you my processing job for your both imagery thermal and visual which I believe that is entirely satisfactory. If it is okay for you, please share it here (at least some screenshots).

 

@channel:

We currently have difficulties in extracting the r-jpeg imagery from FLIR cameras. We have contracted FLIR from our side to see what causes this issue. We believe that the latest changes of FLIR SDK do not align with our algorithms.

As soon I will have news from my side I will comment here.

Hi, as Ina has mentioned she was successful in processing the complete data set. 100% of the images were aligned and calibrated. Took me about 10mins to get the index map from the data once I had those as pix4d is really user friendly. 

I believe the massive difference in how I was processing it when it failed, to a successful output was the following settings were changed- 

Internal parameters optimisation - all prior 

Previous it was set to all. Changing it to all prior changed the calibration to 100% success. 

The other main settings that I have managed to source straight from FLIR support are the following - 

Visible Camera
Sensor width (mm) = 7.4mm
Sensor Height (mm) = 5.55mm
Pixel size (um) = 1.85um
Focal length (mm) = 8mm
Principle point x (mm) = 7.4mm/2 = 3.7mm
Principle point y (mm) = 5.55mm/2 = 2.7525mm

 
Thermal camera
Sensor width (mm) = 640*17um * 10^(-6)m/um = 10.88mm
Sensor Height (mm) = 512*17um * 10^(-6)m/um = 8.704mm
Pixel size (um) = 17um
Focal length (mm) = whatever camera you have (13mm in this case)
Principle point x (mm) = halfway in between the sensor 10.88mm/2= 5.44mm
Principle point y (mm) = halfway in between the sensor 8.704mm/2= 4.352mm

Below is a screen shot of the processed raycloud and calibrated cameras. 

 

Here is the digital surface model… This was all straight from the low resolution Tiff files (640x512) and I still feel the quality we gained is amazing. 

 

 

If anyone needs any other help regarding this you can ask me but Ina is definitely a wizard who can better assist. If I can I will try. 

thanks

1 Like

Hi Ina, 

Just with that example we processed, any idea why the site is somewhat curved south? The site is actually dead straight east to west, but the model is warped as you can see. Not sure why this happened? It’s also really flat with no ridge down the center.

 

 

You also mention an issue extracting data from the radiometric jpeg images, would you recommend we capture the imagery separately as Tiff files and Jpg visible images straight away?

The camera is capable of doing this and it seems to work for the processing, however the export gives weird temperature values, they are not recorded in Degrees celsius. I believe there is an equation to derive the celsius temperatures, but haven’t managed to do it accurately as yet. Taken directly from the duo pro r manual === 

 

The 14-bit TIFF file format of the Duo Pro R camera contains temperature information in the form of pixel intensities. These pixel values can be converted into temperature using the following formulas:

[counts@High Resolution] * 0.04 – 273.15 = deg C

[counts@Low Resolution] * 0.40 – 273.15 = deg C

where “counts@High Resolution” are the individual pixel values recorded using the Temp Range setting of High (-25 to +135C) and “counts@Low Resolution” are pixel values recorded at a Temp Range setting of Low (-40 to +550C)

Hi Mark,

 

The bending it is usually present in projects that are above flat terrain and low in texture. At the moment I am not at my PC to take a look at the data, but once I am back, I will look again. Do you have the, bending in both RGB and thermal? Having GCP’s should correct this behavior.

The problem of extracting will be self-evident since the software will not be able to read the imagery. But in your case, since it does the firmware that you have on your camera most probably does not affect this. We are still investigating this.

As you have an R version of the camera, the temperature should be absolute. However, this depends on which units have been shot initially.

If you can allow me a bit of time, I will look closely to your data these days and come back to you.

 

Best,

Ina

Hi Mark,

 

As expected the thermal project presents the bending while the visual RGB not. This is due to the low feature present in the thermal imagery and other difficulties that I have mentioned in the post before. 

 

However, if you are planning to merge the projects remember that you have the option to use the point cloud from RGB and texture from thermal.

Best,

Ina 

We are finally going forward with the Sensor purchase, Since our main objective is to use the Flir duo pro r with Pix 4dCapture or djiGSpro to collect and pix4dMapper to process. At this time is it possible to plan flights using Pix4dcapture and the Flir duo pro r and a m600or m200 (grid flight plans)

If not what are the thermal duo sensors am i able of using with Pix4d Capture and mapper preferably using DJI drones like m200 or m600? The DJI zen muse XTII

Pedro Penedos

Hi Pedro,

The Flir Duo Pro R works well in mapper but not as a rig. Some tests have been done by our team and the camera are not synchronized well enough. We have no feedback about the Zenmuse XT2.

Also, we did not test it but it should be possible to use a Flir Duo Pro camera on a DJI M600 or DJI M200 by implementing it as a custom camera on the iOS version of Pix4Dcapture. You can find how to define a custom camera in Pix4Dcapture here. The important step is to define the time-lapse directly in the settings of the camera, and then apply the same value for the Min. triggering interval [s] in the application. If not, the camera will not take any picture as it cannot be controlled by Pix4Dcapture.

The camera Zenmuse XT2 is not supported by Pix4Dcapture and we are not planning to implement it in our software at the moment. However, the latest versions of Pix4Dcapture are integrating the DJI SDK 4.6 which introduce support for XT2 camera.
It might work if you use the XT2 camera with these Pix4Dcapture versions but we can not make any guarantee.

About processing thermal images, you can find more information on our knowledge base:
https://support.pix4d.com/hc/en-us/articles/360000173463-Processing-thermal-images
[https://support.pix4d.com/hc/en-us/articles/115005852003-Processing-camera-rigs

](https://support.pix4d.com/hc/en-us/articles/115005852003-Processing-camera-rigs)Best,

1 Like

Dear Gael,

Thanks for the update! Would it be possible to forward the information to Flir about your experience that “the camera are not synchronized well enough”? I think FLIR is also interested in providing greater added value to their products.

Best,

 

New to this thread, but I’m having many of the same issues processing thermal imagery from the FLIR Duo Pro R.  I’m using exactly the parameters and settings specified by Mark Craig on August 16, above, but am still getting garbage results (wildly distorted orthomap, for one).

In my particular case, my UAV flew at a nearly constant 122 meters above the surface in a rectangular grid over a 100 acre field. There’s 361 individual image files, each one nadir-viewing and with reasonably good GPS geolocation.  The surface had some gentle rolling hills, but nothing that should pose any problems, especially when viewing at nadir from that altitude. 

I sincerely believe that if I were able to extract the thermal images individually and print them on paper (I can’t, because the R_JPEG file format is apparently a trade secret), I would have no problem laying them all out on a table and producing a paper mosaic that is “good enough” for my purposes.  Is there a way to tell Pix4Dmapper to do the equivalent in software – trust the geolocation and camera orientation, overlay the images without distortion correction and create a low-precision but reasonable digital mosaic?

Alternatively, what camera internal parameters should I consider adjusting experimentally to see what makes Pix4Dmapper happier?

1 Like

Hi Grant,

We are currently processing your images. We will get back to you soon and will also update in this post.

We were talking with pix4d about this during intergeo’18: we do not need 3D models just the orthophoto. Apparently pix4d fields is able to do this but the two products mapper and fields are licensed separately, and this feature is not included in mapper…

@Gyula

Pix4Dfields does not support thermal imagery at this time.

But you see my point about pix4d making a product tailored to the needs of an application area, I hope! :wink:

We are really trying to consider all user needs and feature requests!

@Grant The best result (though only 19% calibrated images) were possible with:

  1. Use image scale 1
  2. Use geometrically verified matching
  3. Calibration method standard (use this only if the alternative does not work). Alternative calibration assumes that the images are nadir and will fail to calibrate the images if they are oblique.
  4. Use all prior in internal cameras optimization

The images from dual pro R must be processed in two separate projects (thermal and RGB). Pix4D does the splitting of images. The thermal and RGB images the should be separately used in 2 projects (and then merged).

Since thermal images have a very low resolution, it is difficult to extract features and more so when standard calibration has to be used. Thus the images should have at least 90% overlap.